IFR 1900 Self Test Fails

This forum exists for the purposes for discussing service monitors (This includes but is not limited to Motorola, HP, Aeroflex, GD, etc). Additional topics allowed include test procedures, interpretation of test results, where to find information about specific tests, antenna VSWR, return loss testing, duplexer and filter alignment, etc.

Moderator: Queue Moderator

Post Reply
User avatar
Big BOB
Posts: 188
Joined: Tue Sep 04, 2001 4:00 pm

IFR 1900 Self Test Fails

Post by Big BOB »

I have an IFR 1900 that seems to work perfectly, but I have 3 sections that fail Self Test. Power Meter, AM Modulation and FM Deviation. I use these regularly and have no problems. I have even checked with another service monitor to compare readings. However I would like it to pass its tests. Someone told me that a calibration might resolve the problem. Does this sound reasonable?
jry
Posts: 481
Joined: Fri Feb 06, 2004 3:14 pm

Re: IFR 1900 Self Test Fails

Post by jry »

make sure you are getting 0dm out of the T/R connector when running test 16.

depending on how far off from 0dm and which direction calibration may correct.
jry
Posts: 481
Joined: Fri Feb 06, 2004 3:14 pm

Re: IFR 1900 Self Test Fails

Post by jry »

The tolerance on the 0dbm output for test 16 is +- 2dbm.